philips: Problems with 5V supply + Amplitude value (Vpp) solved

ID: 651073
philips: Problems with 5V supply + Amplitude value (Vpp) solved 
04.Mar.24 22:33
369


After years of being sidetracked, I continued the research and repaired the unit successfully.

I replaced IC 424 which was a 74LS05N.

To check all remaining (digital) IC's in the same section, with the power to the PM5134 turned off (!!), I took a digital multimeter in diode-test mode, and measured all pins in succession with the black lead, while the red lead was against the GND pin.

There should be no short (unless of course the connection is explicitly tied to GND by the PCB trace).

This was a very effective measurement method to find defective ICs with internal shorts. This measurement can be carried out on the digital IC's in-circuit ! So, no need for desoldering!

While the 7805 regulator was removed, I used a 5 VDC laboratory power supply, connected to point K.The current measured with an analog mA-meter on opened solder bridge E was about 225mA.

The unit worked fine during an hour or so.

After soldering the original 7805 regulator back in its place, the display unit dimmed immediately after power on, and the mA-meter over solder bridge E indicated 0 current.

So I installed a new NEC 7805 instead of the original 7805, and everything was fine.

Finally I researched the problem of the incorrect Vpp amplitude measurements.

The display indicated only 0.06 or 0.07 at 2 Vpp output. First suspect was the amplitude display conditioner (see Fig.42 Unit 1 - part 2; bottom left) around IC211 (MC1458) on UNIT1, which gets its input via point 115 from the tandem AMPLITUDE potmeter.

Measurements on resistor 563 (right in front of IC211) showed a constant negative voltage of about -15V on both sides of the resistor, irrespective of the position of the AMPLITUDE knob. The same value appeared at other places of the voltage divider resistors.

IC211 is physically located somewhere on the very front of UNIT1.

To cut a long story short: this was caused by a defective trimmer 558 measuring over 4.4M rather than the nominal value of 1K. Trimmer 558 was listed in the components list as 2K2, but I replaced it with a 1K value as was originally present in my unit.

After this, the AMPLITUDE display was reflecting the correct values.

I am finishing by rerunning all the adjustments of the PM5134 as indicated in the manual section 4.3.4. : "Table of checks and adjustments". However I noticed the rotating switches 86, 84, 85 and 87 need also cleaning. As they are consisting of sealed sections, this will be a time-consuming task.